Measurement of entrance skin doses for patients undergoing diagnostic X-ray examination in federal medical centre (F.M.C.) hospital, Azare, Nigeria
DOI:
https://doi.org/10.70882/p5xjej11Keywords:
Entrance skin dose X-ray Thermoluminescence dosimeter Exposure factorsAbstract
The risk associated with exposure to ionizing radiation is dependent on the characteris
tics of the exposed individual. The size and the structure of the individual influence the
absorbed dose distribution in the organs. Although many studies have been conducted on
radiation dose from medical exposure, in Bauchi, there is not much data on the assessment
of patient entrance skin dose (ESD) and the health risk from conventional radiography in
daily clinical practice. In this work, we estimate the ESD for adult patients undergoing
X-ray examination in F.M.C. Hospital, Azare. A total of 40 patients were included in this
study. ESDs were measured using thermoluminescence dosimeter badges attached to the
skin surface of the X-ray beam during chest PA. Pelvic AP, lumbosacral AP, and lumbosacral
lateral examinations. Patients’ biometric variables (such as age, weight, height, sex, and
thickness), X-ray tube parameters (e.g., tube current and tube voltage), exposure time,
focus to skin distance, and focus to film distance were recorded. The results showed that
the mean values for chest PA, pelvic AP, lumbosacral AP, and lumbosacral lateral were
0.28, 2.71, 3.72, and 7.62 mGy, respectively. The measured ESD for this study ranges from
0.18 to 16.49 mGy. These values were found to be in agreement with the guidance level
set by International Atomic Energy Agency and the standard diagnostic reference level.